Safety & Compliance Forum

Safety & Compliance Info & Resources for
The Power Electronics Industry.

Safety & Compliance Database






For copies of Standards, please contact the appropriate agency. PSMA does not provide copies of standards.
Agency (an Agency can be National, International, or a Non-Government Organization or NGO):
CSA Group
CAN/CSA-C22.2 NO. 61010-1-12 (R2017) - Start year: 2012
Location: Americas - Canada

Description:

This is the harmonized CSA, ISA, and UL standard for Electrical Equipment for Measurement, Control, and Laboratory Use. It is the third edition of CAN/CSA-C22.2 No. 61010-1, the third edition of ANSI/ISA-61010-1 (82.02.01), and the third edition of UL 61010-1. This edition of CAN/CSA-C22.2 No. 61010-1 supersedes the previous edition published as CAN/CSA-C22.2 No. 61010-1 in 2004. This edition of ANSI/ISA-61010-1 (82.02.01) supersedes the previous edition of ISA-82.02.01 published in 2004. This edition of UL 61010-1 will supersede the second edition of UL 61010-1 published in 2004. This Standard is based on IEC 61010-1, third edition.

Notes:

This part of IEC 61010 specifies general safety requirements for the following types of electrical equipment and their accessories, wherever they are intended to be used.

a) Electrical test and measurement equipment

This is equipment which by electromagnetic means tests, measures, indicates or records one or more electrical or physical quantities, also non-measuring equipment such as signal generators, measurement standards, power supplies for laboratory use, transducers, transmitters, etc.

NOTE 1 This includes bench-top power supplies intended to aid a testing or measuring operation on another piece of equipment. Power supplies intended to power equipment are within the scope of IEC 61558 (see 1.1.2 h)).

This Standard also applies to test equipment integrated into manufacturing processes and intended for testing manufactured devices.

NOTE 2 Manufacturing test equipment is likely to be installed adjacent to and interconnected with industrial machinery in this application.

b) Electrical industrial process-control equipment

This is equipment which controls one or more output quantities to specific values, with each value determined by manual setting, by local or remote programming, or by one or more input variables.

c) Electrical laboratory equipment

This is equipment which measures, indicates, monitors, inspects or analyses materials, or is used to prepare materials, and includes in vitro diagnostic (IVD) equipment.

This equipment may also be used in areas other than laboratories; examples include self-test IVD equipment to be used in the home and inspection equipment to be used to check people or material during transportation.

1.1.2
Equipment excluded from scope

This Standard does not apply to equipment within the scope of:

a) IEC 60065 (Audio, video and similar electronic apparatus);
b) IEC 60204 (Safety of machinery - Electrical equipment of machines);
c) IEC 60335 (Household and similar electrical appliances);
d) IEC 60364 (Electrical installations of buildings);
e) IEC 60439 (Low-voltage switchgear and controlgear assemblies);
f) IEC 60601 (Medical electrical equipment);
g) IEC 60950 (Information technology equipment including electrical business equipment, except as specified in 1.1.3);
h) IEC 61558 (Power transformers, power supply units and similar);
i) IEC 61010-031 (Hand-held probe assemblies);
j) IEC 61243-3 (Live working - Voltage detectors - Part 3: Two-pole low-voltage type).

1.1.3
Computing equipment

This Standard applies only to computers, processors, etc. which form part of equipment within the scope of this standard or are designed for use exclusively with the equipment.

NOTE Computing devices and similar equipment within the scope of IEC 60950 and conforming to its requirements are considered to be suitable for use with equipment within the scope of this standard. However, some of the requirements of IEC 60950for resistance to moisture and liquids are less stringent than those in this standard (see 5.4.4 second paragraph)).

1.1.4
DV DR Addition of the following referencing the National Electrical Code and the Canadian Electrical Code:

This Standard applies to equipment:

a) To be employed in accordance with ANSI/NFPA 70, National Electrical Code® (NEC);
b) Designed to comply with the general requirements of CAN/CSA C22.2 No. 0 and to be installed in accordance with the Canadian Electrical Code (CEC), Part I, CSA C22.1; or
c) Both (a) and (b).

1.2
Object

1.2.1
Aspects included in scope

The purpose of the requirements of this standard is to ensure that HAZARDS to the OPERATOR and the surrounding area are reduced to a tolerable level. Requirements for protection against particular types of HAZARD are given in Clauses 6 to 13, as follows:

a) electric shock or burn (see Clause 6);
b) mechanical HAZARDS (see Clauses 7 and 8);
c) spread of fire from the equipment (see Clause 9);
d) excessive temperature (see Clause 10);
e) effects of fluids and fluid pressure (see Clause 11);
f) effects of radiation, including lasers sources, and sonic and ultrasonic pressure (see Clause 12);
g) liberated gases, explosion and implosion (see Clause 13).
Requirements for protection against HAZARDS arising from REASONABLY FORESEEABLE MISUSE and ergonomic factors are specified in Clause 16.

RISK assessment for HAZARDS or environments not fully covered above is specified in Clause 17.

NOTE Attention is drawn to the existence of additional requirements regarding the health and safety of labour forces.

1.2.2
Aspects excluded from scope

This Standard does not cover:

a) reliable function, performance, or other properties of the equipment not related to safety;
b) effectiveness of transport packaging;
c) EMC requirements (see the IEC 61326 series);
d) protective measures for explosive atmospheres (see the IEC 60079 series).

1.3
Verification

This Standard also specifies methods of verifying that the equipment meets the requirements of this standard, through inspection, TYPE TESTS, ROUTINE TESTS, and RISK assessment.

1.4
Environmental conditions

1.4.1
Normal environmental conditions

This Standard applies to equipment designed to be safe at least under the following conditions:

a) indoor use;
b) altitude up to 2 000 m;
c) temperature 5 °C to 40 °C;
d) maximum relative humidity 80 % for temperatures up to 31 °C decreasing linearly to 50 % relative humidity at 40 °C;
e) MAINS supply voltage fluctuations up to ±10 % of the nominal voltage;
f) TRANSIENT OVERVOLTAGES up to the levels of OVERVOLTAGE CATEGORY II;
NOTE 1 These levels of transient overvoltage are typical for equipment supplied from the building wiring.
g) TEMPORARY OVERVOLTAGES occurring on the MAINS supply.
h) applicable POLLUTION DEGREE of the intended environment (POLLUTION DEGREE 2 in most cases).

NOTE 2 Manufacturers may specify more restricted environmental conditions for operation; nevertheless the equipment must be safe within these normal environmental conditions.

1.4.2
Extended environmental conditions

This Standard applies to equipment designed to be safe not only in the environmental conditions specified in 1.4.1, but also in any of the following conditions as RATED by the manufacturer of the equipment:

a) outdoor use;
b) altitude above 2 000 m;
c) ambient temperatures below 5 °C or above 40 °C;
d) relative humidity above the levels specified in 1.4.1;
e) MAINS supply voltage fluctuations exceeding ±10 % of the nominal voltage;
f) WET LOCATION;
g) TRANSIENT OVERVOLTAGES up to the levels of OVERVOLTAGE CATEGORY III or IV (see Annex K).

  Sort by Start year / Status
Agency (an Agency can be National, International, or a Non-Government Organization or NGO):
International Electrotechnical Commission
IEC 61010-1-3:2010 - Start year: 2010
Location: Global - Global

Description:

IEC 61010-1:2010 specifies general safety requirements for the following types of electrical equipment and their accessories, wherever they are intended to be used.
a) Electrical test and measurement equipment
b) Electrical industrial process-control equipment c) Electrical laboratory equipment
This third edition cancels and replaces the second edition published in 2001. It constitutes a technical revision.

Notes:

-The scope of the standard has been expanded to include all locations where these products may be used, so that both professional and non-professional versions of these products are within the scope.
- The requirements for testing and measuring circuits (in various subclauses and the entirety of Clause 16) have been removed and included in a particular standard IEC 61010-2-030.
- Insulation requirements (6.7) have been completely rewritten.
- Additional requirements for protection against mechanical HAZARDS (Clause 7) have been included.
- Surface temperature limits (Clause 10) have been modified to conform to the limits of EN 563.
- Radiation requirements (Clause 12) have been modified, and take into account a distinction between intended emission and unintended emission.
- Requirements for reasonably foreseeable misuse and ergonomic aspects have been added (Clause 16).
- A new clause (Clause 17) has been added to deal with HAZARDS and environments not covered by the standard, along with a new informative annex (Annex J) dealing with RISK assessment.
- A new informative annex (Annex E) addresses methods of reducing the POLLUTION DEGREE of a micro-environment.
- Requirements for the qualification of coatings for protection against POLLUTION have been added (Annex H).
- A new informative annex (Annex I) has been added to further explain how to determine the WORKING VOLTAGE of a MAINS CIRCUIT.
It has the status of a group safety publication, as specified in IEC Guide 104. The contents of the corrigendum 1 of May 2011, the interpretation sheet 1 of February 2013 and the corrigendum 2 of October 2013 have been included in this copy.

  Sort by Start year / Status
Agency (an Agency can be National, International, or a Non-Government Organization or NGO):
International Electrotechnical Commission
IEC 61010-1:2010 - Start year: 2010
Location: Global - Global

Description:

IEC 61010-1:2010 specifies general safety requirements for the following types of electrical equipment and their accessories, wherever they are intended to be used.
a) Electrical test and measurement equipment
b) Electrical industrial process-control equipment c) Electrical laboratory equipment
This third edition cancels and replaces the second edition published in 2001. It constitutes a technical revision. This edition includes the following significant changes from the second edition, as well as numerous other changes.
- The scope of the standard has been expanded to include all locations where these products may be used, so that both professional and non-professional versions of these products are within the scope.
- The requirements for testing and measuring circuits (in various subclauses and the entirety of Clause 16) have been removed and included in a particular standard IEC 61010-2-030.
- Insulation requirements (6.7) have been completely rewritten.
- Additional requirements for protection against mechanical HAZARDS (Clause 7) have been included.
- Surface temperature limits (Clause 10) have been modified to conform to the limits of EN 563.
- Radiation requirements (Clause 12) have been modified, and take into account a distinction between intended emission and unintended emission.
- Requirements for reasonably foreseeable misuse and ergonomic aspects have been added (Clause 16).
- A new clause (Clause 17) has been added to deal with HAZARDS and environments not covered by the standard, along with a new informative annex (Annex J) dealing with RISK assessment.
- A new informative annex (Annex E) addresses methods of reducing the POLLUTION DEGREE of a micro-environment.
- Requirements for the qualification of coatings for protection against POLLUTION have been added (Annex H).
- A new informative annex (Annex I) has been added to further explain how to determine the WORKING VOLTAGE of a MAINS CIRCUIT.
It has the status of a group safety publication, as specified in IEC Guide 104. The contents of the corrigendum 1 of May 2011, the interpretation sheet 1 of February 2013 and the corrigendum 2 of October 2013 have been included in this copy.

Notes:

This part of IEC 61010 specifies general safety requirements for the following types of electrical equipment and their accessories, wherever they are intended to be used.

a) Electrical test and measurement equipment

This is equipment which by electromagnetic means tests, measures, indicates or records one or more electrical or physical quantities, also non-measuring equipment such as signal generators, measurement standards, power supplies for laboratory use, transducers, transmitters, etc.

NOTE 1 This includes bench-top power supplies intended to aid a testing or measuring operation on another piece of equipment. Power supplies intended to power equipment are within the scope of IEC 61558

(see 1.1.2 h)).This standard also applies to test equipment integrated into manufacturing processes and

intended for testing manufactured devices.

NOTE 2 Manufacturing test equipment is likely to be installed adjacent to and interconnected with industrial

machinery in this application.

b) Electrical industrial process-control equipment

This is equipment which controls one or more output quantities to specific values, with

each value determined by manual setting, by local or remote programming, or by one or

more input variables.

c) Electrical laboratory equipment

This is equipment which measures, indicates, monitors, inspects or analyses materials, or

is used to prepare materials, and includes in vitro diagnostic (IVD) equipment.

This equipment may also be used in areas other than laboratories; examples include self-

test IVD equipment to be used in the home and inspection equipment to be used to check

people or material during transportation.

1.1.2 Equipment excluded from scope

This standard does not apply to equipment within the scope of:

a) IEC 60065 (Audio, video and similar electronic apparatus);

b) IEC 60204 (Safety of machinery – Electrical equipment of machines);

c) IEC 60335 (Household and similar electrical appliances);

d) IEC 60364 (Electrical installations of buildings);

e) IEC 60439 (Low-voltage switchgear and controlgear assemblies);

f) IEC 60601 (Medical electrical equipment);

g) IEC 60950 (Information technology equipment including electrical business equipment,

except as specified in 1.1.3);

h) IEC 61558

(Power transformers, power supply units and similar);

i) IEC 61010-031 (Hand-held probe assemblies);

j) IEC 61243-3 (Live working – Voltage detectors – Part 3: Two-pole low-voltage type).

1.1.3 Computing equipment

This standard applies only to computers, processors, etc. which form part of equipment within the scope of this standard or are designed for use exclusively with the equipment.

NOTE Computing devices and similar equipment within the scope of IEC 60950 and conforming to its requirements are considered to be suitable for use with equipment within the scope of this standard. However, some of the requirements of IEC 60950 for resistance to moisture and liquids are less stringent than those in this standard (see

5.4.4 second paragraph)).

1.2 Object

1.2.1 Aspects included in scope

The purpose of the requirements of this standard is to ensure that HAZARDS to the OPERATOR and the surrounding area are reduced to a tolerable level.

Requirements for protection against particular types of HAZARD are given in Clauses 6 to13, as follows:

a) electric shock or burn (see Clause 6);

b) mechanical HAZARDS (see Clauses 7 and 8);

c) spread of fire from the equipment (see Clause 9);

d) excessive temperature (see Clause 10);

e) effects of fluids and fluid pressure (see Clause 11);

f) effects of radiation, including lasers sources, and sonic and ultrasonic pressure (see

Clause 12);

g) liberated gases, explosion and implosion (see Clause13).

Requirements for protection against HAZARDS arising from

REASONABLY FORESEEABLE MISUSE and ergonomic factors are specified in Clause16.

RISK assessment for HAZARDS or environments not fully covered above is specified in Clause 17.

NOTE Attention is drawn to the existence of additional requirements regarding the health and safety of labour forces.

1.2.2

Aspects excluded from scope

This standard does not cover:

a) reliable function, performance, or other properties of the equipment not related to safety;

b) effectiveness of transport packaging;

c) EMC requirements (see the IEC 61326 series);

d) protective measures for explosive atmospheres (see the IEC 60079 series).

1.3 Verification

This standard also specifies methods of verifying that the equipment meets the requirements

of this standard, through inspection,

TYPE TESTS,

ROUTINE TESTS, and RISK assessment.

.4 Environmental conditions

1.4.1 Normal environmental conditions

This standard applies to equipment designed to be safe at least under the following conditions:

a) indoor use;

b) altitude up to 2 000 m;

c) temperature 5 °C to 40 °C;

d) maximum relative humidity 80 % for temperatures up to 31 °C decreasing linearly to 50 %

relative humidity at 40 °C;

e) MAINS supply voltage fluctuations up to ±10 % of the nominal voltage;

f)TRANSIENT OVERVOLTAGES up to the levels of OVERVOLTAGE CATEGORY ll;

NOTE 1 These levels of transient overvoltage are typical for equipment supplied from the building wiring.

g) TEMPORARY OVERVOLTAGES occurring on the MAINS supply.

h) applicable POLLUTION DEGREE of the intended environment (POLLUTION DEGREE 2 in most cases).

NOTE 2 Manufacturers may specify more restricted environmental conditions for operation; nevertheless the equipment must be safe within these normal environmental conditions.

1.4.2

Extended environmental conditions

This standard applies to equipment designed to be safe not only in the environmental conditions specified in

1.4.1, but also in any of the following conditions as RATED by the manufacturer of the equipment:

a) outdoor use;

b) altitude above 2 000 m;

c) ambient temperatures below 5 °C or above 40 °C;

d) relative humidity above the levels specified in

1.4.1;

e)MAINS supply voltage fluctuations exceeding ±10 % of the nominal voltage;

f)WET LOCATION;

g) TRANSIENT OVERVOLTAGES up to the levels of OVERVOLTAGE CATEGORY III or IV (see Annex K).

  Sort by Start year / Status
Agency (an Agency can be National, International, or a Non-Government Organization or NGO):
Underwriters Laboratories
UL 61010-1:2012 - Start year: 2012
Location: Global - Global

Description:

This part of IEC 61010 specifies general safety requirements for the following types of electrical equipment and their accessories, wherever they are intended to be used.

This is equipment which by electromagnetic means tests, measures, indicates or records one or more electrical or physical quantities, also non-measuring equipment such as signal generators, measurement standards, power supplies for laboratory use, transducers, transmitters, etc. NOTE 1 This includes bench-top power supplies intended to aid a testing or measuring operation on another piece of equipment. Power supplies intended to power equipment are within the scope of IEC 61558 (see 1.1.2 h)).

This standard also applies to test equipment integrated into manufacturing processes and intended for testing manufactured devices. NOTE 2 Manufacturing test equipment is likely to be installed adjacent to and interconnected with industrial machinery in this application.

Notes:

This is equipment which controls one or more output quantities to specific values, with each value determined by manual setting, by local or remote programming, or by one or more input variables.

This is equipment which measures, indicates, monitors, inspects or analyses materials, or is used to prepare materials, and includes in vitro diagnostic (IVD) equipment.

This equipment may also be used in areas other than laboratories; examples include self-test IVD equipment to be used in the home and inspection equipment to be used to check people or material during transportation.

  Sort by Start year / Status
Agency (an Agency can be National, International, or a Non-Government Organization or NGO):
Underwriters Laboratories
UL 61010-2-030:2012 - Start year: 2012
Location: Global - Global

Description:

Safety requirements for electrical equipment for measurement, control, and laboratory use – Part 2-030: Particular requirements for testing and measuring circuits

Notes:

Equipment included in scope

This part of IEC 61010 specifies safety requirements for testing and measuring circuits which are connected for test or measurement purposes to devices or circuits outside the measurement equipment itself.

These include measurement circuits which are part of electrical test and measurement equipment, laboratory equipment, or process control equipment. The existence of these circuits in equipment requires additional protective means between the circuit and an OPERATOR.

NOTE 1 These testing and measuring circuits may, for example:

- measure voltages in circuits of other equipment,

- measure temperature of a separate device via a thermocouple,

- measure force on a separate device via a strain gauge,

- inject a voltage onto a circuit to analyze a new design

NOTE 2 Testing and measuring circuits that are not within the scope of this Part 2 are considered to be covered by the requirements of Part 1

NOTE 3 Equipment containing these testing and measuring circuits may be intended for performing tests and measurements on hazardous conductors, including MAINS conductors and telecommunication network conductors. See Annex BB for considerations of HAZARDS involved in various tests and measurements

a) Electrical test and measurement equipment

This is equipment which by electromagnetic means tests, measures, indicates or records one or more electrical or physical quantities, also non-measuring equipment such as signal generators, measurement standards, power supplies for laboratory use, transducers, transmitters, etc.

NOTE 1 This includes bench-top power supplies intended to aid a testing or measuring operation on another piece of equipment. Power supplies intended to power equipment are within the scope of IEC 61558 (see 1.1.2 h)).

This standard also applies to test equipment integrated into manufacturing processes and intended for testing manufactured devices.

NOTE 2 Manufacturing test equipment is likely to be installed adjacent to and interconnected with industrial machinery in this application.

b) Electrical industrial process-control equipment

This is equipment which controls one or more output quantities to specific values, with each value determined by manual setting, by local or remote programming, or by one or more input variables.

c) Electrical laboratory equipment

This is equipment which measures, indicates, monitors, inspects or analyses materials, or is used to prepare materials, and includes in vitro diagnostic (IVD) equipment. This equipment may also be used in areas other than laboratories; examples include self-test IVD equipment to be used in the home and inspection equipment to be used to check people or material during transportation.

Equipment excluded from scope

This standard does not apply to equipment within the scope of:

a) IEC 60065 (Audio, video and similar electronic apparatus);

b) IEC 60204 (Safety of machinery - Electrical equipment of machines);

c) IEC 60335 (Household and similar electrical appliances);

d) IEC 60364 (Electrical installations of buildings);

e) IEC 60439 (Low-voltage switchgear and controlgear assemblies);

f) IEC 60601 (Medical electrical equipment);

g) IEC 60950 (Information technology equipment including electrical business equipment, except as specified in 1.1.3);

h) IEC 61558 (Power transformers, power supply units and similar);

i) IEC 61010-031 (Hand-held probe assemblies);

j) IEC 61243-3 (Live working - Voltage detectors - Part 3: Two-pole low-voltage type).

Computing equipment

This standard applies only to computers, processors, etc. which form part of equipment within the scope of this standard or are designed for use exclusively with the equipment.

  Sort by Start year / Status
Agency (an Agency can be National, International, or a Non-Government Organization or NGO):
Underwriters Laboratories
UL 61010-1:2019 - Start year: 2019
Location: Global - Global

Description:

Safety Requirements for Electrical Equipment for Measurement, Control, and Laboratory Use - Part 1: General Requirements

Notes:

This revision to ANSI/UL 61010-1 dated July 19, 2019 is an editorial correction to Table I.1, per IEC 61010-1, Cor1:2019 to align with the IEC version.

1.1.1 Equipment included in scope

This group safety publication is primarily intended to be used as a product safety standard for the products mentioned in the scope, but shall also be used by technical committees in the preparation of their publications for products similar to those mentioned in the scope of this standard, in accordance with the principles laid down in IEC Guide 104 and ISO/IEC Guide 51.

This part of IEC 61010 specifies general safety requirements for the following types of electrical equipment and their accessories, wherever they are intended to be used.

a) Electrical test and measurement equipment

This is equipment which by electromagnetic means tests, measures, indicates or records one or more electrical or physical quantities, also non-measuring equipment such as signal generators, measurement standards, power supplies for laboratory use, transducers, transmitters, etc. NOTE 1 This includes bench-top power supplies intended to aid a testing or measuring operation on another piece of equipment. Power supplies intended to power equipment are within the scope of IEC 61558 (see 1.1.2 h)).

This standard also applies to test equipment integrated into manufacturing processes and intended for testing manufactured devices. NOTE 2 Manufacturing test equipment is likely to be installed adjacent to and interconnected with industrial machinery in this application.

b) Electrical industrial process-control equipment

This is equipment which controls one or more output quantities to specific values, with each value determined by manual setting, by local or remote programming, or by one or more input variables.

c) Electrical laboratory equipment

This is equipment which measures, indicates, monitors, inspects or analyses materials, or is used to prepare materials, and includes in vitro diagnostic (IVD) equipment.

This equipment may also be used in areas other than laboratories; examples include self-test IVD equipment to be used in the home and inspection equipment to be used to check people or material during transportation.

1.1.2 Equipment excluded from scope

This standard does not apply to equipment within the scope of:

a)    IEC 60065 (Audio, video and similar electronic apparatus);

b)    IEC 60204 (Safety of machinery - Electrical equipment of machines);

c)    IEC 60335 (Household and similar electrical appliances);

d)    IEC 60364 (Electrical installations of buildings);

e)    IEC 60439 (Low-voltage switchgear and controlgear assemblies);

f)    IEC 60601 (Medical electrical equipment);

g)    IEC 60950 (Information technology equipment including electrical business equipment, except as specified in 1.1.3);

h)    IEC 61558 (Power transformers, power supply units and similar);

i)    IEC 61010-031 (Hand-held probe assemblies);

j)    IEC 61243-3 (Live working - Voltage detectors - Part 3: Two-pole low-voltage type).

1.1.3 Computing equipment

This standard applies only to computers, processors, etc. which form part of equipment within the scope of this standard or are designed for use exclusively with the equipment.

NOTE Computing devices and similar equipment within the scope of IEC 60950 and conforming to its requirements are considered to be suitable for use with equipment within the scope of this standard. However, some of the requirements of IEC 60950for resistance to moisture and liquids are less stringent than those in this standard (see 5.4.4 second paragraph)).

1.1.4DV Addition of the following referencing the National Electrical Code and the Canadian Electrical Code:

This standard applies to equipment:

a)    To be employed in accordance with ANSI/NFPA 70,  National Electrical Code® (NEC);

b)    Designed to comply with the general requirements of CAN/CSA C22.2 No. 0 and to be installed in accordance with the Canadian Electrical Code (CEC), Part I, CSA C22.1; or

c)    Both (a) and (b).

1.2 Object

1.2.1 Aspects included in scope

The purpose of the requirements of this standard is to ensure that HAZARDS to the OPERATOR and the surrounding area are reduced to a tolerable level.

Requirements for protection against particular types of HAZARD are given in Clauses 6 to 13, as follows:

a)    electric shock or burn (see Clause 6);

b)    mechanical HAZARDS (see Clauses 7 and 8);

c)    spread of fire from the equipment (see Clause 9);

d)    excessive temperature (see Clause 10);

e)    effects of fluids and fluid pressure (see Clause 11);

f)    effects of radiation, including lasers sources, and sonic and ultrasonic pressure (see Clause 12);

g)    liberated gases, explosion and implosion (see Clause 13).

Requirements for protection against HAZARDS arising from REASONABLY FORESEEABLE MISUSE and ergonomic factors are specified in Clause 16.

RISK assessment for HAZARDS or environments not fully covered above is specified in Clause 17.

NOTE Attention is drawn to the existence of additional requirements regarding the health and safety of labour forces.

1.2.2 Aspects excluded from scope

This standard does not cover:

a) reliable function, performance, or other properties of the equipment not related to safety;

b) effectiveness of transport packaging;

c) EMC requirements (see the IEC 61326 series);

d) protective measures for explosive atmospheres (see the IEC 60079 series).

1.3 Verification

This standard also specifies methods of verifying that the equipment meets the requirements of this standard, through inspection, TYPE TESTS, ROUTINE TESTS, and RISK assessment.

1.4 Environmental conditions

1.4.1 Normal environmental conditions

This standard applies to equipment designed to be safe at least under the following conditions:

a) indoor use;

b) altitude up to 2 000 m;

c) temperature 5 °C to 40 °C;

d) maximum relative humidity 80 % for temperatures up to 31 °C decreasing linearly to 50 % relative humidity at 40 °C;

e) MAINS supply voltage fluctuations up to ±10 % of the nominal voltage;

f) TRANSIENT OVERVOLTAGES up to the levels of OVERVOLTAGE CATEGORY II;

NOTE 1 These levels of transient overvoltage are typical for equipment supplied from the building wiring.

g) TEMPORARY OVERVOLTAGES occurring on the MAINS supply.

h) applicable POLLUTION DEGREE of the intended environment (POLLUTION DEGREE 2 in most cases).

NOTE 2 Manufacturers may specify more restricted environmental conditions for operation; nevertheless the equipment must be safe within these normal environmental conditions.

1.4.2 Extended environmental conditions

This standard applies to equipment designed to be safe not only in the environmental conditions specified in 1.4.1, but also in any of the following conditions as RATED by the manufacturer of the equipment:

a) outdoor use;

b) altitude above 2 000 m;

c) ambient temperatures below 5 °C or above 40 °C;

d) relative humidity above the levels specified in 1.4.1;

e) MAINS supply voltage fluctuations exceeding ±10 % of the nominal voltage;

f) WET LOCATION;

g) TRANSIENT OVERVOLTAGES up to the levels of OVERVOLTAGE CATEGORY III or IV (see Annex k).

  Sort by Start year / Status
We welcome your input!
Use this button to let us know about any standards, agencies or other
information you believe we've left out, or any errors you find. Thank you.